Multiple Choice
An automated blood gas analyzer performs two-point calibrations that match expected values.Subsequent quality control runs show unacceptable values for PO2 and PCO2,but pH is acceptable.Which of the following is the most likely problem?
A) Contaminated buffers
B) Contaminated calibration gas
C) Analyzer temperature not at 37° C
D) Computer error
Correct Answer:

Verified
Correct Answer:
Verified