Short Answer
What is the approximate entrance skin exposure for a 20-cm AP pelvis produced at 80 kVp, 20 mA, and a SID of 40 in (100 cm), if 70 kVp = 2.5 mR/mA?
Correct Answer:

Verified
Correct Answer:
Verified