Multiple Choice
A new high frequency radiographic system is being released to a customer for first clinical use.The applications person states that the AEC density controls are in 20% increments.If a normal AEC exposure of the hip is 24 mAs, a repeat exposure at +2 would yield a mAs value of
A) 26 mAs.
B) 29 mAs.
C) 34 mAs.
D) 48 mAs.
Correct Answer:

Verified
Correct Answer:
Verified