Short Answer
A satisfactory radiograph is produced using the fixed kVp theory.The part measuring 10 cm and 8 mAs was used.To maintain exposure, what mAs should be used if the part measures 14 cm?
Correct Answer:

Verified
Correct Answer:
Verified