Multiple Choice
If the linear attenuation coefficients for bone and water are 0.380 and 0.190, respectively, and the scaling factor of the scanner is 1000, the computed tomography number for bone is:
A) +1000
B) 1000
C) +380
D) +190
Correct Answer:

Verified
Correct Answer:
Verified