Multiple Choice
A radioactive source of 137Cs (t1/2= 30 y) was calibrated on October 23, 2000, to contain 10 Ci.This is used as a daily accuracy check source in the dose calibrator.Presuming the dose calibrator is working properly, what activity should the dose calibrator show on April 23, 2006?
A) 9.6 Ci
B) 8.8 Ci
C) 5.4 Ci
D) 2.2 Ci
Correct Answer: