Multiple Choice
While analyzing a blood gas control sample, you notice a trending of several PCO2 values above the 2 SD range over the last six control samples run. What corrective action would you consider at this time?
I.Check the function of the PCO2 electrode.
II.Rerun the control sample a second time.
III.Repair or replace any failed components.
A) I and II
B) I and III
C) II and III
D) I, II, and III
Correct Answer:

Verified
Correct Answer:
Verified