Multiple Choice
You are called by a nursing home to help set up O2 for a patient, delivered through a jet nebulizer. Unfortunately, all that is available is an old uncompensated Thorpe tube flowmeter. When set to 8 L/min, the nebulizer creates an additional 30 psig of downstream pressure. Approximately what flow of O2 will the patient now receive?
A) 4 L/min
B) 6 L/min
C) 8 L/min
D) 12 L/min
Correct Answer:

Verified
Correct Answer:
Verified