Multiple Choice
A patient is being treated on the 18-MV beam to a PA spine at a depth of 10 cm.The collimator setting is 10 × 13 with no blocks.The dose is 300 cGy per fraction.What is the dose delivered to Dmax?
A) 375 cGy
B) 240 cGy
C) 400 cGy
D) 300 cGy
Correct Answer:

Verified
Correct Answer:
Verified